I have the OSTC Mk2; having just upgraded to latest stable v3.10 and setting auto set point change auto. When I manually change from set point 1 to set point 2 in the simulator then the diluent automatically changes to 32/0. I then have to change the diluent back to air throgh the menu page.
My set up is as follows, all diluents set to air. Diluent 1 set active.
Gas menu is gas 1,3,4 and 5 set to air; gas 2 set to 40/0. Gas 1 is first gas active, all other gasses set not active.
Is this a bug or intended operation and if intended why?
Thanks in advance
John
Manual setpoint change problem
Re: Manual setpoint change problem
Additional info:
I re-flashed my OSTC with v3.02 software and the situation as desribed in post 1 does not happen; ie, when I manually change from set point 1 to 2 and 3 the diluent remains the same and does not auto change to 32/0.
Regards
John
I re-flashed my OSTC with v3.02 software and the situation as desribed in post 1 does not happen; ie, when I manually change from set point 1 to 2 and 3 the diluent remains the same and does not auto change to 32/0.
Regards
John
-
- Posts: 4464
- Joined: 13 May 2007, 18:07
Re: Manual setpoint change problem
Hi John,
Thanks for reporting - I'll investigate this issue tomorrow.
Regards,
Matthias
Thanks for reporting - I'll investigate this issue tomorrow.
Regards,
Matthias
-
- Posts: 4464
- Joined: 13 May 2007, 18:07
Re: Manual setpoint change problem
Hi,
I can confirm this bug. It's indeed a serious issue for CCR divers (Well, at least you can see that the wrong diluent is chosen).
Sorry for that.
I'll upload the 3.11 with this issue fixed today.
regards,
Matthias
I can confirm this bug. It's indeed a serious issue for CCR divers (Well, at least you can see that the wrong diluent is chosen).
Sorry for that.
I'll upload the 3.11 with this issue fixed today.
regards,
Matthias
-
- Posts: 4464
- Joined: 13 May 2007, 18:07
Re: Manual setpoint change problem
Hi,
Have some issues with our firmware section on the webpage currently, but here's the 3.11: read.php?2,15834 And it's also available via the "Web Download" in the Tinybootloader v2.
Regards,
Matthias
Have some issues with our firmware section on the webpage currently, but here's the 3.11: read.php?2,15834 And it's also available via the "Web Download" in the Tinybootloader v2.
Regards,
Matthias
Re: Manual setpoint change problem
Matthias,
I can confirm that software v3.11 no longer has the issues described at post 1. Many thanks for the quick fix.
Regards
John
I can confirm that software v3.11 no longer has the issues described at post 1. Many thanks for the quick fix.
Regards
John